Post by Freseniusguy123 on Mar 1, 2024 22:45:19 GMT
Fresenius slow flow calibrations have given me issues. Every machine I’ve updated has come to the back with high temp issues and replacing all 3 temp sensors has solved my issues and recalibrating. I’ve had to do condo cell re-calibrations also. Sometimes the machines condo and temp are perfect yet still need to replace sensors or recalibrate. Anyone else running into this?
|
|
Icebonez
Junior Member
What is better - to be born good, or to overcome your evil nature through great effort?
Posts: 55
|
Post by Icebonez on Mar 4, 2024 13:09:34 GMT
Is the "Temp Comp" setting enabled in service mode?
|
|
|
Post by freseniusguy123 on Mar 4, 2024 20:55:15 GMT
Yes. They required us to activate it on all the machines. I've had people in my area complain about these issues. I'm just confused why the slow flow calibration is causing so much problems.
|
|
|
Post by thor on Mar 4, 2024 21:28:28 GMT
When the machine drops down to 100 flow, After test with no PT connected or TX started have seen temp increase and condo sometimes get out of limits. Once staff puts in the prescribed flow rate, etc. Temp and condo should stabilize within 3-5 minutes and remain stable. The 100 flow rate is to save concentrates when HD is sitting "idle" without Tx running, etc.
|
|
|
Post by Freseniusguy123 on Mar 18, 2024 2:11:05 GMT
To Thor. I agree on that my machines randomly clime up to 40. I’ve had 60% of my machines who got through the update so that on the floor. I end up telling them to pull it and replace all the temp sensors. Recalibrate everything. When I let the machines that climb up in temp randomly than go back down end up causing issues later on. Now I’m just replacing them all during my PM’s. Not worth the issues later on.
|
|