Post by FloridaMan24 on Mar 2, 2024 0:27:57 GMT
Hey all, New to the forum and completely new to Braun machines. I had a tech pull a machine today for blood detected in Prep/disinfection. How would I go about figuring out what the issue is? I'm running it now in dialysis therapy and the machine has passed all tests without issues? Any ideas would much appreciated!
|
|
|
Post by Chuck Weddle on Mar 2, 2024 11:12:54 GMT
There have been issues with SOME Streamlines being just a little too tinted and the SAD thinks it sees blood. If it starts happening on other machines too this is probably the problem.
That said, SAD's do go bad.... Try calibrating the SAD. If the problem comes back, replace it.
|
|
|
Post by Admin on Mar 2, 2024 16:33:21 GMT
You can verify if this is the issue by moving the streamline tubing “up higher on the door” so to speak to get a different “area” of tubing if you feel it’s cloudy looking.
|
|
|
Post by FloridaMan24 on Mar 4, 2024 1:15:58 GMT
Thanks for info!
|
|