Post by Biomed06 on Mar 13, 2024 15:51:46 GMT
I have a k2 machine 30,000 hours having intermittent flow inlet error. I checked all the possible causes and basically replaced almost all hydraulic components, I'll say the hydraulic system is new. I'm at lost , the error keeps popping out. I noticed after letting it ran and get the system warmed up , the error will go away and passes test but when you run it the following day it error will show again.
|
|
|
Post by Guest2day on Mar 13, 2024 16:07:13 GMT
Would you mind listing all components you have replaced in the hydraulics? Did you perform a back to back before and after new install of the parts? A back to back will be the best thing to do at this time to further determine hydraulics or electronics.
Check the boards in the cage & check dist board for burns. Once the parts you replaced or checked are listed, that will help us direct you better. I don't want to give you the general check rundowns b/c I'm sure you've been through it too.
|
|
|
Post by Biomed06 on Mar 13, 2024 16:17:08 GMT
Thanks for responding. I replaced all the valves ( 29,41,30,39,24,25,26) , balance chamber assy with valves , float swtich, dearation and flow pump with heads. cfs 9 and 10, air separator assy. rebuilt inlet water regulator . rebuilts loading and flow pressure regulators. How do you perform back to back ? sorry my first time hearing it.
|
|
|
Post by dave74 on Mar 13, 2024 17:52:18 GMT
The float switch is not lowering the way it should, or else the machine senses it that way. Usually this is caused by a problem with the deaeration pump/motor or an obstruction downstream, but you have replaced almost everything.
I suppose it could be something electrical, and as the machine warms up, the error goes away.
In a back to back, you hook up the machines in such a way that the electronics of one controls the hydraulics of the other. You want to see if the problem follows the card cage or the hydraulics. It can get confusing. Guest2day has a point. I am out of ideas.
|
|
|
Post by Biomed06 on Mar 13, 2024 18:48:45 GMT
Thanks Dave, My focus was on the hydraulic side and then got frustrated with all the parts and the countless hours I spent replacing them so I let the machine untouched for now. I try doing the back to back and go from there.
|
|
pacnw
Full Member
Posts: 187
|
Post by pacnw on Mar 13, 2024 19:24:19 GMT
in the tips and tricks section of this forum is a flow error work sheet. Follow that to find the issue.
|
|
|
Post by Guest2day on Mar 13, 2024 19:36:00 GMT
Thanks for listing what you did as well. Fingers crossed the back to back narrows it down to one of the two.
Double check that your valves are connected to the appropriate spot on the dist board and vice-versa on each valve is connected correctly. Make sure valve #24 & 26 connectors aren't swapped with eacother. If I remember correctly valve 26's wiring pins are soldered differently from all the rest of the valve connections.
If you do end up determining that it's the hydraulics.. go back to checking the BC assembly. Eventhough it's new, check all valves, open them up to check for rusting and make sure there are no obstructions in the valve housing. I've come across very small pieces of broken plastic/debris inside those valve housings. Also, I see you rebuilt the loading valve- if you use diasafe filters it needs to be the strong spring.
|
|
|
Post by Biomed1 on Mar 14, 2024 2:58:40 GMT
Not common but I've seen a diasafe filter cause this. And please ensure your pump heads aren't installed backwards. Input should be lower than the output.
|
|
|
Post by gnurk on Mar 14, 2024 12:40:20 GMT
be advised when 25k hours is reached the y and t connectors can flake off and cause an obstruction they should be replaced at 25k hours
|
|
|
Post by biomed111 on Mar 17, 2024 14:35:42 GMT
I've had an issue similar to this, it turned out the deaeration orifice was intermittently getting blocked by some debris.
|
|
Icebonez
Junior Member
What is better - to be born good, or to overcome your evil nature through great effort?
Posts: 55
|
Post by Icebonez on Mar 18, 2024 12:17:49 GMT
Thanks for responding. I replaced all the valves ( 29,41,30,39,24,25,26) , balance chamber assy with valves , float swtich, dearation and flow pump with heads. cfs 9 and 10, air separator assy. rebuilt inlet water regulator . rebuilts loading and flow pressure regulators. How do you perform back to back ? sorry my first time hearing it. I've seen flow motor brushes cause this. Have you checked/replaced them?
|
|
|
Post by Mph92563 on Mar 18, 2024 21:37:28 GMT
Mine was caused by the Balancing Chamber, crack in the diaphragm.
|
|
|
Post by gnurk on Mar 19, 2024 12:08:02 GMT
also check your act/sensor cables
|
|
|
Post by 48yrshuntinbubbles on Apr 1, 2024 0:17:09 GMT
Back when we had some high hour Ks, we had a few hydro blocks start breaking down inside. We replaced it, problem went away.
|
|