Post by Ranicia on Mar 15, 2024 0:44:16 GMT
How to correct 10 fill program 1
|
|
pacnw
Full Member
Posts: 187
|
Post by pacnw on Mar 15, 2024 14:31:33 GMT
Filling program issues are usually caused by air leaks. Somewhere there is air being introduced into the system. Most common spots are the hansen hoses and flow bob assembly, UF sample port, all tubing on the air seperator and or the air seperator itself... These are all good spots to start.
|
|
|
Post by gnurk on Mar 17, 2024 12:20:23 GMT
read your ts guide under fill programs could be v69 or V43 is not venting
|
|