Post by dialysisbob on Mar 15, 2024 19:38:12 GMT
Hi,
As the thread says.. when doing calibration on both pumps neither is sucking water in the burette and the machine just counting down with with no sucking, also the machine is stop and go with no movement in the burette.
This is a BS, bibag machine with approximately 24k and this is what I did prior. Checked both daeration and flow and has new brushes, CFS 9 and 10, actuator and power logic board, check the BC membranes and tried new acid and bicarb pumps.
Any help is appreciated.
Thanks
|
|
|
Post by Biomed1 on Mar 16, 2024 2:35:41 GMT
Any V104 or V105 errors in dialysis?
Any flow errors in rinse or dialysis mode?
Dearation, loading and flow pressures good?
|
|
|
Post by gnurk on Mar 17, 2024 12:18:35 GMT
did you check your pumps is it just one or both
|
|