Post by Him1342 on Mar 19, 2024 14:47:45 GMT
I have had a machine down for a very long time with a BiBag Temp Sensor Error. And I have also found out that there are other Techs experiencing the same issue without resolution. Here is a list of what has been done/replaced:
Pre-Temperature Sensor
Post-Temperature Sensor
BiBag Temperature Thermistor
BiBag Pressure Transducer
BiBag Hydraulic Assembly
BiBag Interface Board
Mother Board
Function Board
It will go away and reappear the following day or two, without the machine ever being used. Anyone experienced this, and can you help me with this?
Thank you.
|
|
pacnw
Full Member
Posts: 187
|
Post by pacnw on Mar 19, 2024 15:06:06 GMT
Judging from what you have completed already i'd lean towards the Sensor and/or actuator board or cables to the boards. Don't forget t&c mode for the sensor board. Hope this helps. The actuator board sends the signal to the sensor board but the sensor board is what controls the temp....
|
|