2008T, non BiBag, 30K hours After startup, Power Logic Board Fail Replaced Power Logic Board, Actuator, Function, Sensor Boards. One at a time Still shows up on power up. Clears after rinse. Any ideas
|
|
pacnw
Full Member
Posts: 187
|
Post by pacnw on Mar 19, 2024 15:47:11 GMT
Turn the machine off. Take your function board out and carefully remove the yellow battery RAM chip. Pay attention to its orientation so you can put it back in the right way. Leave it disconnected for 2 minutes. Place it back in the function board and reconnect the wires. Boot the machine up in service mode. Re-enter the time and date, machine serial number, machine hours, and your name. Save that info and shut the machine off. Turn the machine on and let it boot up in dialysis. Error should be gone. Now do a FULL rinse. When that's done turn the machine off and unplug from the wall for 2 minutes. Should be good to go now.
|
|
|
Post by middleofnowhere on Mar 20, 2024 13:30:15 GMT
Still present after trying the Yellow Bat Ram reset. Any other ideas??
|
|
pacnw
Full Member
Posts: 187
|
Post by pacnw on Mar 20, 2024 14:33:01 GMT
That's worked for me everytime I see this alert.... Call the help desk or try the steps again... Sorry it didn't work!
|
|
|