Post by kts on Mar 27, 2024 20:58:44 GMT
Good afternoon all, I have a version 9 machine that I get no power at all. I checked the power cord continuity all good switched the SMPS same problem, I changed main switch still nothing I also checked the old switch all was good but it was worth a try. i get nothing no lights no motors no alarms just dead machine. Any Ideas what to look at? also I did check the outlet first, and checked all the fuses.
|
|
|
Post by Chuck Weddle on Mar 27, 2024 21:05:51 GMT
Did you check for voltage at the power cord terminals on the SMPS? Are any LED's on the SMPS lit? Trace the main switch wiring to the SMPS.
|
|
|
Post by SrCusEngr on Mar 27, 2024 21:36:32 GMT
<kts> The mains switch does not control 120 VAC. It only controls +5 VDC on one side and Ground on the other side. Mess that up and it is no-go. Its purpose is to control the go-to-sleep/wake-up circuits in the power supply. With you doing all you said, there are only two possibilities (well 3): no power; mis-wired mains switch; GFCI tripping (no power). If you put in a good SMPS, it has to be either before it or something is telling it to stay off (misplaced ground - switch wiring). All wires look alike but they each serve a different purpose.
|
|
|
Post by kts on Mar 28, 2024 11:53:35 GMT
Did you check for voltage at the power cord terminals on the SMPS? Are any LED's on the SMPS lit? Trace the main switch wiring to the SMPS. Yes I have voltage at SMPS no LEDs are lightning only thing I can think of is maybe I wired the switch wrong but I took off terminal and put on new switch but I will double check that.
|
|
|
Post by kts on Mar 28, 2024 14:47:22 GMT
Thanks for the help it was the switch wired wrong.
|
|
|
Post by SrCusEngr on Mar 28, 2024 15:33:09 GMT
Thanks for posting your reply.
|
|