nbabo
New Member
Posts: 2
|
Post by nbabo on Apr 4, 2024 0:58:43 GMT
Hi all,
software version 9.1E with 9385 Working Hours. Test 1.27 fails at LLC 12550. Is this a leaky Membrane? How do i confirm?
|
|
|
Post by SrCusEngr on Apr 4, 2024 13:29:24 GMT
<nbabo> Fast check: 1. Go to Test 1.13 BC Valves. 2. Take a 60 cc syringe filled with water and connect it to the circuit at FBK1. 3. Open VEBK1/VABK1. 4. Push on the plunger, if it empties the syringe, you have a bad membrane. If it stops with resistance, Chamber 1 is good. Refill and check chamber 2. 5. Open VEBK2/VABK2. 6. Push on the plunger, if it empties the syringe, you have a bad membrane. If it stops with resistance, Chamber 2 is good.
|
|
nbabo
New Member
Posts: 2
|
Post by nbabo on Apr 5, 2024 4:50:30 GMT
Thank You. Another machine has blown Power supply so swapped for this one. Will do the test with Syringe once the power supply arrives.
|
|