Post by Ghost707 on Apr 4, 2024 18:14:34 GMT
so I have a machine that for the past 3 months it has a 105 stuck closed I have replaced all the normal components w/calibrations(also confirming at end of rinse) now its saying 105 stuck closed while patient is running but the machine is not alarming, still pulling uf, condo and temp are fine. just the "stuck closed banner"in red not alarming. I never ran across this before.
|
|
|
Post by 48yrshuntinbubbles on Apr 8, 2024 21:36:38 GMT
Is the machine in bypass? Is Bob going up and down during treatment? If not, you may be pulling weight off, but the diffusion side of the dialysis process isn't happening, you're just "puffing" them. Last time I had this issue, a check valve broke and had sprayed the gen2 module, soaking it and blowing the board inside.
|
|