Post by Timmy on Apr 6, 2024 16:55:44 GMT
How does one change the NSVB in a Dialog plus SWV9.1E?
|
|
|
Post by Admin on Apr 6, 2024 21:57:12 GMT
Disconnect the wiring harness to the float, undo the lid screws, push the pins out of the molex connector, undo the nut, slide the sensor out of the under side of the lid.
|
|
|
Post by SrCusEngr on Apr 8, 2024 21:24:35 GMT
<Timmy> You will be loosening the outside nut. Before using your wrench, mark the threads at and above the nut with a black marker. Why? There is also a nut inside. If that gets moved, it will change your float position for LOW. you now have a guide on the old unit, that you can transfer to the new threads as a sort of gauge to set the right place for the external nut. You will also need a tool to remove the pins. It is found in Extraction Tool Set, 7701756.
|
|
|
Post by Timmy on Apr 26, 2024 17:09:26 GMT
Thanks, just received the extraction kit and the part.
|
|