Post by Deleted on Sept 2, 2021 18:24:35 GMT
Hello everyone,
I am getting a no chemical intake alarm. Any suggestions?
|
|
BMF
Junior Member
Posts: 86
|
Post by BMF on Sept 2, 2021 19:10:11 GMT
Is it taking up any bleach when you try to duplicate the error?
|
|
Deleted
Deleted Member
Posts: 0
|
Post by Deleted on Sept 2, 2021 20:30:45 GMT
Yes it is
|
|
|
Post by g on Sept 2, 2021 20:57:23 GMT
Make sure it's not leaking out usually from hydro-chamber. If not leaking then possible board issue.(actuator)
|
|
stuff
Full Member
1111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111
Posts: 230
|
Post by stuff on Sept 3, 2021 11:35:19 GMT
Bleach is sucked up through the acid wand. Its first stop is the acid pump. Rebuild the pump. Start simple then go complicated with boards. It could also be check valves if you use those
|
|
|
Post by Gambrita on Sept 3, 2021 16:59:08 GMT
Yup, I agree with stuff. Check the acid pump springs are usually the culprit. also look at your wand, uptake filter, check valve if you have one in your acid line.
|
|
|
Post by lisagalvan on Sept 17, 2021 4:54:05 GMT
Have had to rebuild the acid pumps before.And,also if the bleach bottle tubing had tiny hole so it could pull any air,it would give me that alarm.Watch on debug screen,and when conductivity gets around 9,it will start counting.
|
|
|
Post by Henry on Jan 26, 2022 18:49:45 GMT
Hello everyone, I am getting a no chemical intake alarm. Any suggestions? Usually it's air in the system, but I once had the drain valve staying open and hydrochamber was not filling up properly.
|
|
|
Post by tobinfly on Jan 11, 2023 21:04:23 GMT
usually after rebuilding acid pump (2008T machine) the problem should be fixed. But in some cases the machine will suck bleach and then throw the error "No chemical intake" at this point look for a leak in BIBAG AIR SEPARATION CHAMBER.
|
|
|
Post by guest on Jan 12, 2023 16:45:07 GMT
Float switch if the springs don't work.
|
|
quilla
New Member
Life is Good.
Posts: 25
|
Post by quilla on Jan 18, 2023 17:02:51 GMT
Make sure the bleach bottle straw is not loose, for it is loose it wont pull the bleach.
|
|
phillipashleyludlow
Full Member
01001000 01101001 00100000 01010011 01110100 01110101 01100110 01100110 00101110
Posts: 151
|
Post by phillipashleyludlow on Jan 20, 2023 15:13:59 GMT
Bleach is sucked up through the acid wand. Its first stop is the acid pump. Rebuild the pump. Start simple then go complicated with boards. It could also be check valves if you use those yup.
|
|
|
Post by Shecky on Feb 2, 2024 2:52:35 GMT
When a machine draws up bleach it fills the hydrochamber until the float switch recognizes that the machine has drawn an adequate amount of bleach. I think it's 300ml. If the float switch doesn't recognize the hydrochamber has filled enough it will throw this alarm. Rebuild your acid pump and if that doesn't work change your float switch. If that doesn't work, it may be a leak or an electrical problem.
|
|
Icebonez
Junior Member
What is better - to be born good, or to overcome your evil nature through great effort?
Posts: 55
|
Post by Icebonez on Feb 2, 2024 14:55:02 GMT
I had this happen with an improperly installed a diasafe filter. Rookie mistake but it was when I first started as a biomed. In agreement with the other posts here, my fixe is usually the acid pump rebuild.
|
|