Post by waterwrench on Apr 15, 2024 15:22:17 GMT
I replaced a hour meter on a 2008T. Now I get an "art BP no comm" banner. No error number on pump, just the banner. Any suggestions? Thank you
|
|
|
Post by dave74 on Apr 15, 2024 15:39:41 GMT
Is this the same machine that had the Rinse Condition High alarm?
|
|
|
Post by dave74 on Apr 15, 2024 16:11:07 GMT
A logical place to start is reseating all the boards and checking all the cables.
I have seen the functional board cause this problem.
|
|
|
Post by gnurk on Apr 16, 2024 12:09:00 GMT
does this machine have the multicolored ribbon cable going from bp module to funct board
|
|
|
Post by waterwrench on Apr 16, 2024 12:58:18 GMT
This is the same machine that had the High rinse condition. I've swapped out cables, pump, reseated & swapped some boards.swapped out power supply. error won't go away.
|
|
|
Post by t1spike on Apr 16, 2024 15:20:40 GMT
are you sure you have ribbon cable on second top socket on function board?
|
|
Icebonez
Junior Member
What is better - to be born good, or to overcome your evil nature through great effort?
Posts: 55
|
Post by Icebonez on Apr 17, 2024 13:03:06 GMT
For what it is worth, I've seen the LP955 board go bad without it actually displaying an error code on the pump.
|
|
|
Post by waterwrench on Apr 17, 2024 14:02:58 GMT
It was the little ribbon cable between the function board and test board. Thanks to everyone for helping.
|
|
|
Post by dave74 on Apr 17, 2024 14:58:49 GMT
Thanks for posting your fix.
|
|