Dialog +91E just changed level sensor in tank. Now this comes up. Machine was idle for a month waiting for parts. How do I troubleshoot this?
|
|
|
Post by Admin on May 3, 2024 11:31:40 GMT
Hi Tim, When it goes into Solution Intake - do you hear the VD valve open? Or you can open the front door and watch it cycle.
|
|
timmy
New Member
Posts: 13
|
Post by timmy on May 3, 2024 12:12:01 GMT
Yes it cycles but citric acid is not pulled in ⁸
|
|
|
Post by Admin on May 3, 2024 12:54:40 GMT
If it's not pulling in, you might have a bit of citric crystalized in the disinfection port where the tubing connects to
|
|
timmy
New Member
Posts: 13
|
Post by timmy on May 3, 2024 13:14:31 GMT
Got the tube cleared by syringing hot water into metal male fitting during solution intake. Thank you for your help!
|
|
|
Post by SrCusEngr on May 3, 2024 17:20:13 GMT
<timmy> I worry when people use a syringe (with needle to get into the port?) to clear that port. If pushed too far, you will never notice that you have punctured the tubing inside, until it fails. Use the short leg of a small paperclip. Even if it goes in too far, it won't immediately slice through the internal tubing. Also, citric acid can crystalize in the tubing going to the jug, if the machine is not used often.
|
|
|
Post by Admin on May 3, 2024 18:04:41 GMT
SrCusEngr - Usually I'll take a 2 inch or so piece of citric tubing and put it over the luer lock of the syringe (fits perfect) and let it hang and gently push the hot water into the port. No needle for me. I hate needles!
|
|
|
Post by Chuck Weddle on May 3, 2024 18:28:42 GMT
SrCusEngr - Usually I'll take a 2 inch or so piece of citric tubing and put it over the luer lock of the syringe (fits perfect) and let it hang and gently push the hot water into the port. No needle for me. I hate needles!With those tats? ?!!!!!!
|
|
|
Post by Admin on May 3, 2024 18:50:22 GMT
That's a different kind of needle!
|
|