dcvd
New Member
Posts: 22
|
Post by dcvd on May 9, 2024 13:26:04 GMT
When you all measure the conductivity on your machine with an independent meter what is yours all range between the both +/- 0.2 or +/- 0.5 and how about the measured dialysate pH value should it be 6.9 to 7.6. for final dialysate or should be 4.0 -8.0 +/- .30 against the independent meter.
I am using a D6 meter. I am just curious to what you all are going by for the parameters.
|
|
|
Post by Chuck Weddle on May 9, 2024 13:38:18 GMT
By regulation, pH must be 6.9 - 7.6, there is no +/-. We use pH strips from RPC. Part #K100-0117 for normal acid concentrate and K100-0117C for citrate based concentrate.
|
|
dcvd
New Member
Posts: 22
|
Post by dcvd on Jun 28, 2024 17:41:10 GMT
What are your paramenters for final conductivity before putting a patient on.
|
|
|
Post by Chuck Weddle on Jun 28, 2024 18:34:58 GMT
What are your paramenters for final conductivity before putting a patient on. The dialysis machine can be used if: - The reading on the hand-held meter and
the dialysis machine display are between 13.1mS and 14.1mS. - The two readings are within +/- 0.5mS of
each other. If the reading is not within BOTH of the above parameters, the machine must be removed from service for repair.
|
|