the walking biomed
New Member
“Sometimes I Feel Like An Idiot, But I Am An Idiot, So It Kinda Works Out” (Billy Madison)
Posts: 15
|
Post by the walking biomed on May 15, 2024 13:07:04 GMT
Starting this thread one at a time because we dont have enough machines for it to be this backed up..
we will start with the first issue. 9.1B. 20026 hours
This Machine had an air issue yesterday flushed and fixed and sent into service, It came back last night with the following.
-"df flow disturbed" -"would not come up to conductivity with liquid OR powder bicarb"
|
|
the walking biomed
New Member
“Sometimes I Feel Like An Idiot, But I Am An Idiot, So It Kinda Works Out” (Billy Madison)
Posts: 15
|
Post by the walking biomed on May 15, 2024 14:00:50 GMT
this may be worth something im not sure. looked at the conductivity test in tsm and when turning on kp it does not move. i have it at 250 and it is not spinning whatsoever
|
|
|
Post by Admin on May 15, 2024 14:14:55 GMT
Does KP spin by hand?
|
|
the walking biomed
New Member
“Sometimes I Feel Like An Idiot, But I Am An Idiot, So It Kinda Works Out” (Billy Madison)
Posts: 15
|
Post by the walking biomed on May 15, 2024 14:43:25 GMT
|
|
|
Post by Admin on May 15, 2024 17:39:12 GMT
So if you have it in TSM and it's not spinning, you have a problem with the motor, the wire harness, or the PBM
|
|
|
Post by Admin on May 15, 2024 17:40:27 GMT
swap electrical connections with the KP and the UFP and turn on the KP pump in TSM... does UFP spin? If yes it's the KP motor. If no, then harness or PBM
|
|