Post by carl on May 29, 2024 5:20:09 GMT
During the solution uptake cycle the machine alarms solution intake failure. I discovered that valve VD is not being opened when I open the wrench/service screen during citric thermal solution uptake. Only valve VZ is. I can go into TSM mode into lower manual menu and command valve VD to open and close. So I know it’s apparently working. Could there be a condition not being met that is causing this to happen, bad cable or connection? Bad small disinfection circuit board? Any help would be greatly appreciated. Machine is Dialog + Evolution and software version is 9.12.
|
|
btech
Junior Member
Posts: 55
|
Post by btech on May 29, 2024 8:21:47 GMT
Hi carl the machine reacts to PDA, which should be more negativ than -100mmHg at this point. So it might be that your PDA is defect/ not tight or your UFP is not getting the Sensor down because of a leak. Lot at PDA during the suck in phase.
|
|