Post by moonbean on Jun 20, 2024 14:05:24 GMT
I have two 2008T machines with 15k+ hours giving “bicarb cond 2 low” these are blue star premium. I’ve Checked my dear/loading and flow pressure, calibrated acid/bicarb pumps and done condo cell calibration. Does anyone have any recommendations?
|
|
BMF
Junior Member
Posts: 96
|
Post by BMF on Jun 20, 2024 16:56:46 GMT
Calibrate your bicarb conductivity cell.
|
|
|
Post by caferacer87 on Jul 2, 2024 21:05:38 GMT
while in dialysis mode go to screen 15 and look at your TCB, your BIC Mon Cond should be with in 10 of the TCB, if its not you have to adjust your loading valve in dialysis mode. Turn slow just incase you over shoot
|
|