Post by bsdreamer on Jun 24, 2024 13:05:06 GMT
I have a bluestar machine (10,000 hrs) that will only let the staff touch the buttons on the very edge not in the middle of the buttons of the treatment screen, is there a way to calibrate only the tx side screen? The chairside screen is working perfect.
|
|
|
Post by stuff on Jun 24, 2024 13:44:10 GMT
Different companies use different chairside programs. This is rough without knowing which program.
For Darwin Direct
1. Log into Darwin Direct or have a floor staff do this. You can do this with a patient running 2. Up at the top in the middle is an icon of a wrench, click that 3. Select Calibrate Screen and follow the instructions.
If it says unable to calibrate then that is a touch screen driver issue and complicates things.
|
|
BMF
Junior Member
Posts: 96
|
Post by BMF on Jun 24, 2024 14:25:28 GMT
Since you said it's a Bluestar machine, you can boot the machine in service mode. The very first prompt once it has booted states to press New TX to calibrate monitor. Touch the dots that appear and it will calibrate the monitor.
|
|
|
Post by stuff on Jun 25, 2024 13:51:23 GMT
Since you said it's a Bluestar machine, you can boot the machine in service mode. The very first prompt once it has booted states to press New TX to calibrate monitor. Touch the dots that appear and it will calibrate the monitor. ^^ This. I read the post as the other way around, sorry
|
|
|
Post by bsdreamer on Jun 27, 2024 12:14:20 GMT
Thanks everyone, I had forgotten about that first screen just used to going straight to the buttons on bottom. It did fix the issue.
|
|
|
Post by dave74 on Jun 27, 2024 14:15:36 GMT
Thanks for posting your fix.
|
|