Post by ksilcox11 on Jul 2, 2024 17:08:18 GMT
Good day all,
I have an 8.3 machine that gives me a temp not reached during heat disinfect the temp goes up to 95 but then drops pretty quick. tried to do a temp calibration and now i cant get the temp to come up at all.
|
|
|
Post by SrCusEngr on Jul 2, 2024 17:16:10 GMT
<ksilcox11> Check F5/F6 fuses in the SMPS. Check the heater rod resistance. R=E2/V (assuming USA) = 14,400/1800 = 8 ohms. If you read an open, you may have blown the internal safety thermal fuse. That would mean you will have to replace the heater rod. NOTE: Never apply power without water flowing thru the heater rod canister.
|
|
|
Post by ksilcox11 on Jul 3, 2024 12:56:14 GMT
I have already checked the fuses and the heater thanks for the suggestion though its something to do with the flows because when I go to start the calibration the pumps will start but stop almost as fast as they start.
|
|
btech
Junior Member
![*](//storage.proboards.com/forum/images/stars/star.png)
Posts: 55
|
Post by btech on Jul 15, 2024 6:28:16 GMT
<ksilcox11> Check F5/F6 fuses in the SMPS. Check the heater rod resistance. R=E 2/V (assuming USA) = 14,400/1800 = 8 ohms. If you read an open, you may have blown the internal safety thermal fuse. That would mean you will have to replace the heater rod. NOTE: Never apply power without water flowing thru the heater rod canister. close, as the heater has two elements with each 900W the resistant is 16 ohm, to be exact 15.4 ohms +-5%
|
|
|
Post by SrCusEngr on Jul 15, 2024 17:45:23 GMT
<btech> Sorry, you are incorrect. This is a Version 8 machine, Dialog+. Even though it is a Dialog+, the Version 8 software means it has a single element heater rod of 1800 W. Only Dialog+ V-9.xx machines have a dual rod (900 W each) for the heater unit
|
|
|
Post by ksilcox11 on Jul 27, 2024 13:13:52 GMT
Found the problem the inlet water valve wasn't opening causing the pumps to shut down replaced the valve and everything calibrated as usual. thanks for the suggestions guys sorry about the late response also I'm only in this clinic once a month.
|
|
|
Post by Chuck Weddle on Jul 27, 2024 14:03:22 GMT
Found the problem the inlet water valve wasn't opening causing the pumps to shut down replaced the valve and everything calibrated as usual. thanks for the suggestions guys sorry about the late response also I'm only in this clinic once a month. A "Water Supply Disturbed" alarm wasn't any kind of indication to you?
|
|
btech
Junior Member
![*](//storage.proboards.com/forum/images/stars/star.png)
Posts: 55
|
Post by btech on Aug 19, 2024 11:58:32 GMT
Yes you are right forgot that its a version 8. <btech> Sorry, you are incorrect. This is a Version 8 machine, Dialog+. Even though it is a Dialog+, the Version 8 software means it has a single element heater rod of 1800 W. Only Dialog+ V-9.xx machines have a dual rod (900 W each) for the heater unit
|
|