Post by biomed111 on Jul 3, 2024 13:18:27 GMT
Hi guys, I don't really know where to start except that this machine has 20k hours and sometimes it will give a 24v low error that will not clear. I've seen it happen during heat but never personally during tx. The 2 times I did see it, i replaced one bibag kit at a time (first the left side then the middle kit). Both times the error cleared but I cannot seem to find anything useful in the troubleshooting guide.
|
|
|
Post by The Bluestar Whisperer on Jul 3, 2024 13:58:28 GMT
Try changing out the 2 capacitors in the power supply. Just had 2 machines at one of my clinics have this come up at around 27K hours that I fixed with this.
|
|
|
Post by biomed111 on Jul 3, 2024 13:59:13 GMT
Also i'd like the mention that I did use the voltage extender board and checked 12v+ and 5v+ and those were fine.
|
|
|
Post by 48yrshuntinbubbles on Jul 4, 2024 12:19:53 GMT
When I have intermittent 24v alarms I blow the dust out of the power supply and cardcage. I also reseat all boards and cables including on the distribution board, that usually rectifies the issue.
|
|