Post by walker on Jul 8, 2024 18:08:01 GMT
Installed new level detector censors. Put it on dialysis mode and set blue plastic to calibration mode. No matter how much I turn the pots, the lights don't light up on the board. Put water in tube and in level detector. When I put it in between the censors, I can press reset button and clear the air detector error. Any ideas on why I can't calibrate?
|
|
|
Post by The Bluestar Whisperer on Jul 8, 2024 18:48:31 GMT
I'll assume that the blue plastic is the service jumper. Were you following the calibration procedure 3.2? What caused you to change out the sensors in the first place?
|
|
|
Post by sheepdog1 on Jul 9, 2024 14:47:46 GMT
Im guessing the board is the original problem not sensors. When the board is bad it wont let you cal the level detector
|
|
Icebonez
Junior Member
What is better - to be born good, or to overcome your evil nature through great effort?
Posts: 72
|
Post by Icebonez on Jul 10, 2024 12:32:06 GMT
It could also be you need to replace your "dummy" drip chamber. Try making a fresh one from new lines and see if that helps?
|
|