Post by biomedbran on Jul 10, 2024 12:30:24 GMT
Hey folks,
Any ideas on why I cant calibrate my balance chamber? It will not prime and says failure to fill.
Thanks so much!
|
|
|
Post by The Bluestar Whisperer on Jul 10, 2024 14:15:06 GMT
You should still be able to perform the calibration, just turn prime back off and proceed as normal. I would do the procedure at least twice to ensure you are getting the correct result.
|
|
|
Post by Phillip-Ashley Ludlow on Jul 10, 2024 14:38:07 GMT
you could try to let it come into cond in dialysis mode b4 you do the calibration in service mode.
just to satiate my own curiosity, why are you calibrateing the BC values?
|
|