Post by guessed on Jul 18, 2024 14:56:02 GMT
when I power on a T machine, rinse is my only option, with a banner on the left side stating: BRAM lost power-up Flags! Lost SVS/UF settings! Lost Dial settings! Could problem be my RAM chip or functional board? I reset all boards, ribbons and cables. When rinsed is pressed, it is in a mandatory rinse. If I power off, then same banner and mandatory rinse. After the rinse, I am able to enter dialysis mode. ![](//storage.proboards.com/7226994/thumbnailer/IgaOcoZxhnXoozsPUuzx.png)
|
|
|
Post by The Bluestar Whisperer on Jul 18, 2024 15:03:42 GMT
- Replace the yellow RAM Batt on the function board - Then let it go through full Mand rinse - Turn machine off - Turn machine on and bring into conductivity - Run both tests - Run diasafe test - Turn machine off - Turn on in service mode - Set time - Turn machine off - Turn machine back on
All of your errors should be gone.
**If you change the whole function board, make sure to remove the old EEPROM and place it onto the new function board or you'll be ready to punch yourself 5 minute later.
|
|
|
Post by guessed on Jul 18, 2024 15:16:57 GMT
Thanks for your input, but I want to clarify what you posted. I replace the yellow RAM on the functional board, but in your last sentence you wrote to remove old eeprom and place it onto the new function board.
Am I replacing both functional board and the yellow RAM?
|
|
|
Post by The Bluestar Whisperer on Jul 18, 2024 15:21:42 GMT
Thanks for your input, but I want to clarify what you posted. I replace the yellow RAM on the functional board, but in your last sentence you wrote to remove old eeprom and place it onto the new function board. Am I replacing both functional board and the yellow RAM? You shouldn't have to, but I felt I should mention it in the event that you do. I edited the last part to clarify.
|
|