Post by h2otech on Jul 30, 2024 11:43:54 GMT
For as long as I can remember, we have used the HACH bicarb pH Reagent strips. They are no longer available. I did find Serim Guardian Bicarb pH II strips, but they are nearly as much per bottle as we were paying for a box of 5 bottles. Two questions... What is everyone else using and when testing bicarb for pH do the strips have to be specific for dialysate?
|
|
|
Post by Chuck Weddle on Jul 30, 2024 12:37:23 GMT
Do you mean the strips for pH testing that has to be done before starting each treatment of for testing a mixed batch of bicarb?
|
|
|
Post by h2otech on Jul 30, 2024 12:44:35 GMT
We have a pH/Conductivity meter for machine testing. We use the strips to ensure that the bicarb clear from our distribution loop.
|
|
|
Post by Chuck Weddle on Jul 30, 2024 13:18:05 GMT
Give Adrian Bachelier at RPC a call....He will point you to the correct test strip.
1-800-647-3873
|
|
|
Post by h2otech on Jul 30, 2024 13:33:09 GMT
Give Adrian Bachelier at RPC a call....He will point you to the correct test strip. 1-800-647-3873 Thank you!
|
|