Post by ksilcox11 on Sept 27, 2024 12:54:50 GMT
Good day all,
I have a v9 machine giving me a concentrate mix ratio alarm I have tried to recal the mix ratio like the book says I set it to 45 and I am still getting the alarm the ratio is running around 55 on the screen. Im not sure where else to go for this problem. any help is greatly appreciated.
|
|
|
Post by Chuck Weddle on Sept 27, 2024 13:05:55 GMT
Did you try calling tech support or submitting a JIRA request?
|
|
|
Post by ksilcox11 on Sept 27, 2024 13:19:00 GMT
not yet that's my next step was hoping someone had this problem before and had suggestions for the fix
|
|
|
Post by Admin on Sept 27, 2024 13:47:31 GMT
Without knowing temperatures, conductivities, pump speeds, or anything else - it's hard to tell.
Can you please provide this?
|
|
|
Post by joey on Sept 27, 2024 14:53:25 GMT
Do you see air in the wand/tubing? if you suck any air before the cond. sensors it will give you ratio alarms.
|
|
|
Post by SrCusEngr on Sept 30, 2024 16:19:36 GMT
<ksilcox11> There is no value of 45 for a standard CONCENTRATE mixing ratio. If you are using 45 X concentrates, your mixing ratio should be set to 38. (Bicarb 25 or 35 [cartridge]) changed with 9.1E software to 27.3 and 33.7 [cartridge]).
|
|
|