Post by motorsports on Oct 14, 2021 17:46:59 GMT
22 machines this morning had this alarm banner on the home screen when staff powered them on to start the day. When put into rinse the alarm banner went away and never came back on 20 of them. 2 of the machines had to be powered off and back on to clear the alarm banner. No further issues throughout the day. Has anyone seen this before? any good explanations? I haven't asked Fresenius yet.
|
|
|
Post by K on Oct 14, 2021 18:27:55 GMT
Watch that blood pump display. Any E.03,E.03 or E.05 is an error code covered under warranty to have that module replaced
|
|
stuff
Full Member
1111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111
Posts: 230
|
Post by stuff on Oct 14, 2021 18:54:10 GMT
On the next power routine it will be back. It alternates every other power cycle. Had 10 out of 24 NEW machines with this problem out of the box. Warrantied that crap so fast
|
|
|
Post by gnurk on Oct 15, 2021 12:34:04 GMT
query has the blood pump modules been cabled meaning a thin multicolored ribbon cable from the function board to the module will need to be added insure that all your machines have this cable
|
|
|
Post by motorsports on Oct 15, 2021 14:37:47 GMT
query has the blood pump modules been cabled meaning a thin multicolored ribbon cable from the function board to the module will need to be added insure that all your machines have this cable They do have this cable and have from their initial install here.
|
|
|
Post by JMrFr on Nov 11, 2021 13:09:42 GMT
Recently I have this problem on two T-Machines, I called Fresenius Tech support told me to calibrate Arterial Pressure problem was solve.
|
|
|
Post by gnurk on Nov 14, 2021 13:46:21 GMT
please post fix
|
|
|
Post by motorsports on Nov 22, 2021 20:43:59 GMT
This ended up being on the night of integrated heat when all the machines were on. That same night the generator was tested. This somehow left all the machines with the alarm on the banner in the morning.
|
|
mtw2
New Member
Posts: 47
|
Post by mtw2 on Nov 23, 2021 14:58:56 GMT
Weird, but thanks for posting your fix, or the problem that caused it.
|
|
|
Post by herbiehancock on Nov 30, 2021 16:44:41 GMT
i did see this issue when our power went out for a couple seconds and came back on. maybe there was a quick power outage???
|
|
|
Post by Miguel Cortes on Aug 2, 2023 2:39:09 GMT
This happens to me every time i do a full load test on the Generator, usually i just cancel the heat disinfect and restart it and it clears the message, i have also tried letting the machine finish the full cycle of the Heat disinfect to see if it will do anything else but it doesnt, once its done the machine just turns off...
|
|
|
Post by briancanufixthis on Sept 22, 2023 14:21:54 GMT
P.S. Recently found out that T machines do not like when you test the generator. Probably should test generator when clinic is closed and machines are off...
|
|
|
Post by Biomed1 on Nov 2, 2023 23:32:53 GMT
I had this problem with 2008T machine and I calibrate Arterial Pressure and that took care the problem.
|
|
|
Post by wbills on Jan 23, 2024 18:26:11 GMT
I’ve had two machines that had this error but cleared when the machine was restarted. The funny thing is, that on both machines the lcd hot meter was dead. The PCT said he heard a hiss just before the error appeared. I have to believe that either the hour meter shorted and produced the error, or a power spike took out the hour meter and caused the error to appear. Has anyone experienced something similar?
|
|