Post by Cg on Oct 23, 2021 0:35:52 GMT
Have calibrated venous pressure multiple times.I have disinfected machine and return to service.Came back with the problem fails optical detector.Noticed venous was back on 10.Why is venous pressure not staying at zero after I calibrated it and is that responsible for optical detector failing?Thanks!
|
|
|
Post by gnurk on Oct 24, 2021 15:42:58 GMT
could be a bad variable pot on your board calibrate it again, to be sure also check your pot by adjusting it and measuring the resistance
|
|
|
Post by bimitiri on Oct 25, 2021 10:51:59 GMT
Venous pressure is not related to optical sensor ... Check if the venous pressure internal transducer is wet . If optical sensor fail again, it could be faulty sensor .
Check the ribbon cabling if they are inserted properly.
|
|
|
Post by dave74 on Oct 25, 2021 12:51:56 GMT
I believe you are incorrect on this, bimitiri. If the venous pressure is off, the machine will fail optical detector. That is why the machine will fail optical detector if someone hooks up the monitor lines before the machine has completed test.
|
|
BMF
Junior Member
Posts: 86
|
Post by BMF on Oct 25, 2021 13:03:21 GMT
I believe you are incorrect on this, bimitiri. If the venous pressure is off, the machine will optical detector. That is why the machine will fail optical detector if someone hooks up the monitor lines before the machine has completed test. Dave is correct, the venous pressure must be able to reach 0 for the test to pass. It's one of the reasons why we can't test the machine with transducer protectors connected. The machine will have a value other than 0 and fail optical detector. Every once in a while staff complain of this failure. I walk out on the floor and tell them to disconnect the transducer protectors, they pass every time.
|
|
|
Post by gnurk on Oct 26, 2021 12:02:39 GMT
swap out the level detector module with a known good one your optical detector may be bad
|
|
|
Post by Guest1027 on Oct 27, 2021 16:06:39 GMT
Could possibly be bad optical detector or LP board. If the machine is new and still under warranty call FMC for replacement of optical or LP. I had a T machine with 2k hours with this issue and it ended up being the LP board after module and board swap troubleshooting. Also, found that if Ven pressure is +/-20, optical detector will fail.
There was a voltage/wave calibration that was posted on RenalWeb a long time ago but I can't seem to recall how it was done unfortunately. I know it was mainly based on the use of a oscilloscope and if you didn't have one; you could adjust based on voltage. Maybe someone else here can post that up here if they remember it.
|
|
|
Post by gnurk on Oct 27, 2021 17:17:00 GMT
we where going to do that electronic calibration until they found out to get the correct oscilloscope would cost 4K
|
|
|
Post by bimitiri on Oct 28, 2021 10:39:17 GMT
Thank you for the correction ... I learnt something today
|
|
jtech
New Member
Posts: 44
|
Post by jtech on Nov 4, 2021 16:24:35 GMT
LP 450 Board, pn M31451
|
|
|
Post by lisagalvan on Nov 15, 2021 2:03:08 GMT
Thanks everyone for the responses!So far...after the last venous calibration..the machine has not came back.I know it can change at any time,though!Appreciate all your help!
|
|
|
Post by dave74 on Nov 15, 2021 12:16:43 GMT
Thanks for the update.
|
|
|
Post by BLA on Mar 7, 2022 16:32:49 GMT
I CHANGE THE MODULE AND STILL I HAVE OD TEST: NO ART ALARM.....
|
|
|
Post by Biomed 304 on Mar 23, 2022 14:55:54 GMT
BLA calibrate arterial it will fix that OD TEST: NO ART ALARM error every time.
|
|