Full Member
Posts: 235
|
Post by joey on Oct 28, 2021 14:40:32 GMT
Good Morning all! So I had to replace the membranes(failing dfs due to raising PDA during 12500) and there was a tear. After I replaced them, the DFS pressure test failed at 12750.....The pump body hadn't been changed in forever so I just changed it and calibrated and did the simulated therapy and could not get it calibrated right it kept taking too much. So now during 12750 it drops even when clamped and it is a new pump body. So what would cause it to drop at 12750 even when clamped and for there to bee too much fluid to be taken off every treatment? Leaky valve? Thanks
|
|
|
Post by Admin on Oct 28, 2021 14:42:06 GMT
Rinse Bridge or VABK 1/2?
|
|
joey
Full Member
Posts: 235
|
Post by joey on Oct 28, 2021 14:55:24 GMT
Never mind I had to resort to saying "well it wasn't happening before you messed with it" so I went back and somehow some of the screws for the balance chamber weren't tight and there was a little fluid on the tray so I tightened them up and going to go through the cal and the simulated therapy again.
|
|
|
Post by Chuck Weddle on Oct 28, 2021 15:00:46 GMT
In my best Homer Simpson impression........
DOH!
|
|
|
Post by Admin on Oct 28, 2021 15:23:59 GMT
Ahhh there it is! Good job, joey!
|
|
joey
Full Member
Posts: 235
|
Post by joey on Oct 28, 2021 15:50:25 GMT
In my best Homer Simpson impression........ DOH!Hey now we are overdue for our inspection so no DOH's please you might jinx us lol
|
|