Post by B-Boy on Jun 30, 2021 19:11:05 GMT
pressure is not being release at 12650
Membranes are free of rips All valves seem to be working PDA is with in range DDE is within range
Any advice helps.
|
|
|
Post by Admin on Jun 30, 2021 19:24:19 GMT
I would check MSBK 1 or MSBK 2. One of them is not reading.
|
|
|
Post by b-boy on Jun 30, 2021 21:14:56 GMT
BINGO MSBK1 was not responding. ordering the board and coil to be safe
Thank you for posting your fix! Chuck
|
|
|
Post by Admin on Jun 30, 2021 21:55:37 GMT
Perfect! Great job!
|
|
|
Post by VA Biomed on Feb 12, 2024 15:34:53 GMT
hearing grinding noise from middle pump and getting 12650 code. what could it be?
|
|
joey
Full Member
Posts: 227
|
Post by joey on Feb 12, 2024 16:21:09 GMT
hearing grinding noise from middle pump and getting 12650 code. what could it be? Is it moving ? Does it stop now and then ?
|
|
btech
New Member
Posts: 43
|
Post by btech on Feb 26, 2024 7:25:17 GMT
hearing grinding noise from middle pump and getting 12650 code. what could it be? Check for any tubes making contact with the motor if not change it asap, might be the magnet.
|
|