Post by lisagalvan on Jul 2, 2021 11:28:57 GMT
2008T machine goes into self test and passes the tests.But when you press reset button to clear the Test Complete banner, a high conductivity banner appears.The machine is within the set conductivity limits as verified by 90XL.I have reseated boards. Ideas,please?Thanks.
|
|
|
Post by gnurk on Jul 2, 2021 16:03:43 GMT
have you called technical services
|
|
|
Post by lisagalvan on Jul 2, 2021 17:39:46 GMT
No...haven't been back to clinic yet.With the new upgrades,I don't know what to expect,as I have not seen this before.
|
|
pacnw
Full Member
Posts: 187
|
Post by pacnw on Jul 2, 2021 20:17:55 GMT
Is it a bibag machine?? Easy fix if it is
|
|
|
Post by lisagalvan on Jul 2, 2021 20:41:14 GMT
No it is not a Bibag...yet,lol.What is the fix on a bibag,please?
|
|
pacnw
Full Member
Posts: 187
|
Post by pacnw on Jul 4, 2021 18:10:47 GMT
so this error has shown up post software upgrade?? I've also had some weird issues pop up after the upgrade with most being cause by the pressure transducer REV C Try replacing the pressure transducers 9 and 10 with the newest version transducer part number 370506-22. This has solved most of the issues I've encountered.
To correct a high condo banner on a bibag machine you need to calibrate BIC MON to fall between BIC low and BIC HIGH. you do this by adjusting your loading pressure regulator up or down until it matches BIC MON. Then reverify your loading pressure to make sure it's still in range.
|
|
|
Post by lisagalvan on Jul 4, 2021 21:21:21 GMT
Yes...I have had banners I've never seen before..aarrgg.Im not familiar with calibrating the bicarb pump volume on a bibag machine but thought it may be close to a 2TOS machine.I will have to research that procedure!!I will order some new pressure transducers so I will have some on hand.Thank you for the reply!
|
|
|
Post by gnurk on Jul 11, 2021 20:28:29 GMT
pacnw thanks for the info on clearing bibag high condo banners
|
|
|
Post by lisagalvan on Jul 12, 2021 2:55:35 GMT
Replaced the actuator board and banner gone!Thanks for all the help!
|
|
|
Post by gnurk on Jul 12, 2021 16:49:45 GMT
lisa thanks for posting your fix
|
|