Post by kendec65 on Jan 25, 2022 14:33:55 GMT
I just ran across a new problem. I have a BP start/stop stuck red warning. Thoughts on what to do?
|
|
joey
Full Member
Posts: 231
|
Post by joey on Jan 25, 2022 16:39:55 GMT
We have version 8.30 but for us it is usually a loose ribbon cable or a bad front panel board. However I have seen disinfectant run down through a loose seal and cause corrosion of the ribbon cable plug and or plug on board. Now and then I've been able to replace a ribbon cable and/or remove oxidation on plug and it works fine after that. Normally when it happens I move the monitor and if it works again with the monitor in a specific spot it is a loose ribbon cable.
|
|
DTECH
New Member
Posts: 21
|
Post by DTECH on Feb 1, 2022 15:36:08 GMT
@joey same with the units around by me, they spray hexaquart straight on the screen and it damages the board, you will probably need to replace the FPB board - Be careful to check the model of your machine as Bbraun has discontinued the old FPB
|
|
|
Post by Chuck Weddle on Feb 1, 2022 16:45:18 GMT
Also the new FPB's don't come with the hard plastic caps over the switches, you need to reuse them from the old board. The caps are just slightly NOT square so their orientation matters when you install the board.
|
|