Post by Admin on Apr 8, 2022 16:23:14 GMT
TSE is indeed the sensor at the bottom - it's directly out of the heater The new 9.1B/9.1E you have just TSE to calibrate by itself, then you common calibrate TSBIC, TSD, TSDS, TSDE, TSDE_S
|
|
|
Post by SrCusEngr on Apr 14, 2022 18:52:58 GMT
The concentrate pumps do not start spinning until the machine has verified that the cartridge is ready for use (which is why you got the message cartridge connected incorrectly) or the wand is out of the holder after the temperature test. In 9.1B, TSDE must be within 0.6 degrees of TSDE_S; TSD must be within 0.6 degrees of TSD_S and TSBIC must be within 3 degrees of TSD/TSDE.
|
|
|
Post by kolev77 on Apr 18, 2022 8:49:19 GMT
Hello, I think TSDE is 28.8 C because the place of TSDE is between VBP and VDE. Can you see the feedback status of the VBP during this time. If VBP is open that is the reason for TSDE low temperature.
|
|