Post by biobill on Jul 8, 2021 14:08:27 GMT
I have a T machine in rinse mode it runs fine as soon as I put it into dialysis mode the balance chamber doesnt run and it goes into flow error. I am scratching my head . I have changed valve 41 35 36 float switch and checked both pumps ,
|
|
|
Post by Guest on Jul 8, 2021 17:17:23 GMT
Check Pressure transducer 9
|
|
|
Post by gnurk on Jul 11, 2021 20:20:34 GMT
check your CFS switch this is what switches your bal chamber in dialyze
|
|
|
Post by SF BioMed on Jul 27, 2021 20:20:20 GMT
Get your loading pressure gauge, put in the acid port if you get 25 psi the problem is on the secondary side, go to debug screen 6 look for BC switch does it read 897, if it read 897 you have a flow error and pressure is not draining. if so you may want to look at the CFS switch. GO to debug screen 10 look for ACFS look for a min it should drop to zero or close to zero and come straight back up. if it does CFS 9 is good. Get a 60ml syringe attach it to fluid sample port filled with water. go to debug screen zero. make sure valve 30 is blue, push the syringe if you can push all the water then valve 30 is good, if not then its valve 30. this is only some step that may help you out. best of luck
|
|
|
Post by gnurk on Jul 28, 2021 12:49:23 GMT
SF good ts info thanks
|
|