Post by Biomed802 on Apr 21, 2022 14:58:38 GMT
Anyone have a flow inlet error that only occurs right at the end of a heat disinfect and is resolved by flipping up and closing the hanson door?
Anyone see this issue and resolve it? I am stumped.
|
|
|
Post by Sheepdog1 on Apr 21, 2022 17:10:27 GMT
have you checked at anything yet ? pressures, motors ETC....?
|
|
|
Post by GuestingCA on Apr 21, 2022 18:36:20 GMT
Check loading pressure during Heat disinfect. Should be above 15PSI, see the TS guide, section 3; flow errors in clean/disinfect. See pg 170, Clean.
Check air separator for slow leaks as well, once that o-ring is worn there will be leaks. You change it completely, change o-ring or tighten screws down a bit to stop the leak temporarily.
|
|
|
Post by MicroTone on Jun 29, 2022 19:58:04 GMT
Recalibrate loading, flow, deaeration, and inlet water pressures. If that doesn't work check brushes.
|
|
bess
Junior Member
Posts: 59
|
Post by bess on Jun 29, 2022 20:11:26 GMT
Check for worn brushes on the flow motor, CFS starting to fail, or bad float switch
|
|
|
Post by gfreely on Jun 30, 2022 11:02:44 GMT
If you had to flip the door to get it to continue, could a solenoid on 24/25 be sticking when it heats up?
|
|
|
Post by gnurk on Jun 30, 2022 12:11:06 GMT
check v26 also
|
|
|
Post by THOR on Mar 19, 2024 21:30:44 GMT
Thank you everyone. Replacing the float switch fixed my issue. Thank you.
|
|
|