Post by bcjammerx on May 4, 2022 12:57:37 GMT
I posted about a blood leak sensor issue and it reminded me that on one of my machines, both v9.12, during blood leak calibration with the red dialyzer connector in the "blood" solution, my water tank over fills by A LOT. It's not the old "clamp the overflow" issue as these have the updated tanks and it's even coming out the lid. I get red colored water flowing out of the tank. But once the red dialyzer connector is back on the bridge it stops doing this, I was even getting a "high" reading for the level detector.
Still I adjusted the float down but that didn't fix it. It doesn't do it in mock treatments or during disinfect or rinse. Any ideas?
|
|
|
Post by Admin on May 4, 2022 13:09:13 GMT
When you're in the TSM Main Menu, pull the acid and bicarb wands out of the holder and lay them on the base. Then go into the Blood Leak Screen
|
|
|
Post by bcjammerx on May 5, 2022 15:05:29 GMT
Thanks!
|
|