Post by biobill on May 4, 2022 17:08:47 GMT
I just took over a unit with a Better water RO. Did a clean and disinfect yesterday. Everything went smooth. This afternoon the staff call and told me that the RO is alarming water quality . The alarm only lasts for a few seconds, but keeps alarming off and on while the storage tank is filling, They are telling me all the numbers are good on the panel. Any Ideas. I have no other Better water systems anywhere near me.
|
|
joey
Full Member
![*](//storage.proboards.com/forum/images/stars/star.png) ![*](//storage.proboards.com/forum/images/stars/star.png)
Posts: 245
|
Post by joey on May 10, 2022 12:49:41 GMT
I know you said all the numbers are good but did you confirm that? When the alarm goes on and off during filling it could be the rejection/conductivity bouncing around the alarm limit. The might run back and it stops alarming and the numbers look good but then the rejection falls below 90% or something like that. Also is there a level sensor in the tank that is faulty? I don't know that system but those are the things I would check....also temperature. Sometimes cleaning and disinfection can throw your product water quality out of wack especially if membranes are older.
|
|
|
Post by GuestTues on May 10, 2022 17:02:31 GMT
That is normal behavior for the Better water RO system. It works similar to the FR-4, it alarms/beeps during flushing cycle until cond/rej comes down to set limits.
Get a cond meter and verify cond, check set limit range(s).
I would reach out to better water to confirm proper operations and to give you peace of mind.
|
|
|
Post by briancanufixthis on Feb 4, 2023 14:50:49 GMT
Tech support at Better water is great!! He really knows the machines!! He is a little long winded though.
|
|