Post by Amy on May 19, 2022 19:55:54 GMT
Just curious how to manually put the 2008t machines into PHT? When doing annual pm.
|
|
|
Post by t1spike on May 19, 2022 21:29:10 GMT
go into debug screen 1/press 1 on keyboard will start test if passed you wont see anything if failed will get error
|
|
|
Post by Guessed on May 20, 2022 8:36:06 GMT
I have a machine that will fail the PHT while a patient is on, then when machine is pulled and I press 1 on the keyboard in debug, the banner goes real fast and it passes multiple times. But will fail again when a patient is on. Possible dialysate transducer??
|
|
|
Post by gnurk on May 20, 2022 12:31:29 GMT
at 800 flow check that your flow dac is at about 160 with pressure at 35-36 psi this has worked good for me
|
|
|
Post by Guessed on May 23, 2022 14:56:50 GMT
Ok, thanks. So I have checked my machine and at 800 flow the dac read171 so I adjusted dac down to 161. Do you think a to high of a dac would cause the machine to fail pht while a patient was on? Just curious.
|
|
|
Post by GuestMonday on May 23, 2022 20:11:31 GMT
Yes, DAC above 165 will cause PHT to fail when patients are on.
|
|
|
Post by babybiomed92 on Jun 2, 2022 17:16:07 GMT
How do you change the flow DAC? I am having same problem with my online PHT failing and flow pressure is good at 35 but not sure how to adjust the flow DAC you are talking about? I am a new biomed so bare with me here!
|
|
|
Post by GuestHelp on Jun 2, 2022 20:29:23 GMT
To calibrate flow DAC: Make sure to set flow at 800. Turn the nut on the back of the flow pumphead CCW to decrease DAC. Loosen nut at small increments. It may take a few seconds for the DAC value to drop.
|
|
|
Post by babybiomed92 on Jun 10, 2022 14:48:14 GMT
Thank you so very much! Greatly appreciated! It worked out perfect and my PHT passes! thank you!
|
|
|
Post by JazPassingBy on Jan 18, 2024 21:48:08 GMT
After adjusting with in range... and still getting intermittent PHT failures? It'll be stable for a few and will passed... then will fail again and vice versa and so on... What are we looking at, replacing pressure senor/s? And if the machine have bibag system... which sensor is most likely failing, 110? Or perhaps shall we look for something / somewhere else?
|
|