Post by Fresenius123 on Jun 9, 2022 18:59:54 GMT
I have a flow recirc error 1 during heat disinfect. I replaced Dear Motor, & brushes on the flow motor but still getting flow recirc error 1. Troubleshooting guide recommended checks but they all passed them. Could it be a valve issue, heat exchanger or actuator board issue.
Thanks
|
|
|
Post by gfreely on Jun 9, 2022 19:00:52 GMT
I usually see V29 the culprit for that.
|
|
|
Post by Sheepdog1 on Jun 9, 2022 22:30:32 GMT
do you have flow at the flow indicator when you get the error ? check v24 if not
|
|
|
Post by Guest123 on Jun 10, 2022 1:55:23 GMT
|
|
|
Post by gnurk on Jun 10, 2022 12:20:21 GMT
go thru ts guide on flow errors in cleaning programs
|
|
|
Post by Tech on Jun 10, 2022 12:58:32 GMT
its your flow recirc vavle
|
|
|
Post by Fresenius123 on Jun 10, 2022 15:00:05 GMT
I went through the entire troubleshooting guide and I didn't seem to fail any of the test they told me to run. All pressures are normal. NO valve errors. I got this error before when replacing brushes and just moving tubes around deaeration clears it. I'm assuming it might be valve 29 or a kink somewhere in the system.
|
|
|
Post by Sheepdog1 on Jun 10, 2022 16:36:37 GMT
need some more info ... what kind of machine and hours helps and does the error happen when you first put into heat or during the circulation phase ? physically open the valve solenoid see if the plunger assembly is locked up (rusted closed )
|
|
|
Post by Fresenius123 on Jun 10, 2022 18:00:24 GMT
2008T Fresenius machine no bibag, 25,141 for hours. Brushes were done last year but once I got the error I decided to replace the brushes on both motors. It takes about 8-10 seconds before it goes into an error. The remaining rinse time is 0:00 so it doesn't even run a rinse.
|
|
eric
Full Member
Posts: 114
|
Post by eric on Jun 10, 2022 18:24:34 GMT
Motor brushes are seated? motors turning full speed? vent tubing kinked? Deair restrictor open?
|
|
|
Post by Guest123 on Jun 10, 2022 18:58:16 GMT
Are the motors running CCW?
|
|
|
Post by Fresenius123 on Jun 13, 2022 13:13:35 GMT
Its running CCW. All set in place and running. Going to try valve 29 swap today.
|
|
|
Post by Fresenius123 on Jun 13, 2022 14:27:55 GMT
NO luck on the valve 29 or 24 replacements. Any idea where to look for a kink in the system.
|
|
|
Post by Fresenius123 on Jun 13, 2022 15:38:11 GMT
Ended up being the entire flow & deration pump that's both went bad.
|
|
bess
Junior Member
Posts: 59
|
Post by bess on Jun 13, 2022 16:53:11 GMT
Was it the pump heads that were bad or the complete assemblies?
|
|