Post by Chuck Weddle on Jun 27, 2022 15:08:25 GMT
Now we are having issues with lines leaking blood anytime from 40 minutes to an hour and 45 minutes into treatment. One leak was from the pump segment tubing itself very close to the heparin tee. All of the others have been from the heparin tee itself. So far, it has only been Lot# 00VL82280.   
|
|
joey
Full Member
 
Posts: 186
|
Post by joey on Jun 27, 2022 16:04:57 GMT
Man that brings back memories of when Madisystems used to make our non streamline lines. We had a rash of failures exactly like the middle picture.
|
|
|
Post by qman213 on Jun 28, 2022 16:01:42 GMT
We've had a few leak from the pump segment as well. I reported it to BBraun at the time. I honestly assume our staff were getting a little rough with the line when they were stringing them. Maybe not.
|
|