jtech
New Member
Posts: 44
|
Post by jtech on Jul 12, 2022 20:11:18 GMT
I have a Bluestar machine with about 12k hrs giving this error sporadically throughout heat disinfect. I have changed degas and flow pumps, V29 and V30, Diasafe filter, loading pressure regulator, float switch, and reseat the function and actuator boards and it still gives the error. I have ordered some new pressure transducers and going to try changing #9....any other suggestions while I wait??
|
|
|
Post by Sheepdog1 on Jul 13, 2022 2:59:40 GMT
do you have flow at the flow indicator during the error? if not check V24, ohm it out and open the solenoid make sure it not seized up. with that low of hours i would suspect probably not seized but maybe Dead ! ohms are usually around 60
|
|
|
Post by Fresenius123 on Jul 13, 2022 15:12:53 GMT
Had the same thing changing the entire coupling on the Deration motor fixed it. I also had to give it a slight kick nothing to crazy and making sure there are no kinks.
|
|
jtech
New Member
Posts: 44
|
Post by jtech on Jul 14, 2022 14:09:18 GMT
do you have flow at the flow indicator during the error? if not check V24, ohm it out and open the solenoid make sure it not seized up. with that low of hours i would suspect probably not seized but maybe Dead ! ohms are usually around 60 I do have flow during the error. I went ahead and changed V24 anyway and it didn't fix the problem. Gonna try the pressure transducer next. I'm not sure what else to try
|
|
|
Post by Dardee on Jul 14, 2022 18:06:41 GMT
i have had valve #39 cause this by staying stuck closed.
|
|
|
Post by RverdaleTech on Jul 16, 2022 13:08:26 GMT
You may have the flow indicator rise and fall. But how high does it rise? If the bob does not rise above the half-way mark, the problem is with valve 25. Please post your fix.
|
|
jtech
New Member
Posts: 44
|
Post by jtech on Jul 18, 2022 13:26:13 GMT
Changed pressure transducer #9 and it seems to have fixed the issue. Thank you all for your suggestions
|
|
|
Post by gnurk on Jul 19, 2022 11:10:31 GMT
thanks for posting your fix
|
|
|
Post by Fishnqueen56 on Feb 22, 2023 14:39:49 GMT
I had the same problem and changing transducer #10 fixed my problem with flow recirc error. 1
|
|
tom
New Member
Posts: 2
|
Post by tom on Mar 3, 2023 15:21:02 GMT
Try changing part# 647145 {Flow relief valve/Regulator} should solve Flow recirc error
|
|