Post by asmitty on Jul 13, 2021 10:45:28 GMT
I have a Bluestar with 5000hrs. Runs fine 1st shift but once you hit New Tx it starts pouring out the vent tube. At first I thought the Tech just accidently turned off dialysate but its happened 2 more times. if you turn the machine off and back on once the conductivity straightens out everything is fine. Thoughts?
|
|
|
Post by gnurk on Jul 13, 2021 11:00:53 GMT
could be a sticking float switch or heat exchanger call tech services may be a warranty item
|
|
|
Post by asmitty on Jul 15, 2021 11:25:44 GMT
It ended up being the air separator. I couldn't recreate it during tx simulation but it started showing multiple filling programs.
|
|
|
Post by gnurk on Jul 15, 2021 13:13:26 GMT
asmitty thanks for posting your fix
|
|