Post by TX_Stang on Jul 13, 2022 16:59:48 GMT
I have a 2008T machine (10k hours) it got pulled cause upon installing the bibag it’ll burst. It also displayed v104/108 stuck closed I checked all pressures and calibrated the pressure transducers.
Has anyone had this issue before?
Any advise will really help. Thanks in advance
|
|
|
Post by gfreely on Jul 13, 2022 18:50:34 GMT
I *think* one of our guys had this problem and it ended up being V101.
|
|
|
Post by whitechocolate on Jul 13, 2022 18:51:53 GMT
That is two separate issues, both of them use pressure transducer 110. Id suspect transducer 110 is the culprit. I have 8T bibag machines with around the same hours. I cant tell you how many transducers I have replaced in the last year. If it is revision D or older, I'd replace with revision E.
|
|
|
Post by txstang on Jul 13, 2022 21:39:11 GMT
So it’s pressure transducer 110?
|
|
|
Post by stewie on Jul 14, 2022 1:43:43 GMT
I agree about the pressure transducers. I had 15 brand new bibag machines where I had to replace the acid, bibag, dialysate, and CFS pressure transducers in all 15 within the first 8 months or so of having them.
|
|
|
Post by NYPQCTech on Jul 14, 2022 20:57:03 GMT
If you can get the transducers, right now we can't get any spare parts!
|
|
|
Post by anon’s face on Jul 16, 2022 3:15:03 GMT
Had this happen multiple times and at first couldn’t duplicate. It’d pop up occasionally. Only could be 2 things: the valve that allows water into the bag, or the pressure transducer not reading correctly. Wasn’t either, lol. Ended up being the board on the bibag assembly.
|
|
|
Post by bimitiri on Jul 16, 2022 3:57:38 GMT
I had similar issue before and it's pressure transducer 110 . Make sure you don't make the same mistake as I did was to double check that the new replacement pressure transducer is 130mmHg. I accidentally used the Pdial one which was about 1 bar I think. And result was another' kaboom ' and ackward moment.
|
|
|
Post by jaretac on Jul 17, 2022 20:48:40 GMT
I had similar issue before and it's pressure transducer 110 . Make sure you don't make the same mistake as I did was to double check that the new replacement pressure transducer is 130mmHg. I accidentally used the Pdial one which was about 1 bar I think. And result was another' kaboom ' and ackward moment. um, they are all the same pressure transducer. CFS is the only one that is different and that is just because it includes tubing. In fact I just order the 370506-22 for everything including CFS. My guess is you had another bad transducer.
|
|
|
Post by MicroTone on Aug 10, 2022 18:50:09 GMT
If you haven't already fixed, there is a board in the bibag section of the hydraulics that is partially exposed and prone to getting wet. Would be worth checking out. I had the same symptoms once, led me to that board.
|
|
|
Post by gnurk on Aug 11, 2022 12:28:38 GMT
fill out a product complaint or it will not get fixed
|
|