Adnan
New Member
B Braun
Posts: 41
|
Post by Adnan on Jul 14, 2022 12:49:51 GMT
I have a machine with this error ''Bicarbonate Cartidge not connected correctly''. any possible solutions ?
|
|
|
Post by Admin on Jul 14, 2022 13:02:21 GMT
DBK Spring / Disc Blockage at the filter Blockage at the Y Connector Bottom Piercing Spike issue
|
|
|
Post by bimitiri on Jul 16, 2022 10:09:31 GMT
Is one of the hall sensor on the bicarbonate cartridge holder in the right position / loose or faulty ?
|
|
|
Post by bimitiri on Jul 16, 2022 10:19:07 GMT
Sorry i confused with Baxter and Nipro machine earlier, i think Bbraun uses the reel sensor ( Part number : 34517138 ) instead of the hall sensor which located in the lower arm of the bicarbonate cartridge holder . Apparently this part also come with o ring and contacts .
|
|
Adnan
New Member
B Braun
Posts: 41
|
Post by Adnan on Jul 16, 2022 12:12:51 GMT
Is one of the hall sensor on the bicarbonate cartridge holder in the right position / loose or faulty ? Yes it is, you can even check that on TSM mode.
|
|
Adnan
New Member
B Braun
Posts: 41
|
Post by Adnan on Jul 16, 2022 12:15:14 GMT
I will let you know if it works, thanks in advance.
|
|
|
Post by Chuck Weddle on Jul 16, 2022 12:20:45 GMT
That alarm is issued if the machine doesn't see the conductivity spike from priming, not if the Solcart holder is open or not. As Admin stated, DBK is the most common cause. We have also seen kinked tubing from both the top and bottom of the Solcart inside the cabinet.
|
|
Adnan
New Member
B Braun
Posts: 41
|
Post by Adnan on Jul 16, 2022 12:54:38 GMT
That alarm is issued if the machine doesn't see the conductivity spike from priming, not if the Solcart holder is open or not. As Admin stated, DBK is the most common cause. We have also seen kinked tubing from both the top and bottom of the Solcart inside the cabinet. I have never disassembled DBK before, should I open it and clean it?
|
|
|
Post by Chuck Weddle on Jul 16, 2022 15:07:31 GMT
If you open it, I would just replace the disc and spring.
|
|
|
Post by Admin on Jul 16, 2022 16:02:21 GMT
Agreed, open it and replace it. It’s quick and easy. Also pull the Y connector and make sure each side isn’t occluded.
|
|
Adnan
New Member
B Braun
Posts: 41
|
Post by Adnan on Jul 17, 2022 12:58:52 GMT
Agreed, open it and replace it. It’s quick and easy. Also pull the Y connector and make sure each side isn’t occluded. Thanks a lot, it works very well.
|
|
Adnan
New Member
B Braun
Posts: 41
|
Post by Adnan on Jul 17, 2022 12:59:25 GMT
If you open it, I would just replace the disc and spring. Appreciated sir, it works.
|
|
|
Post by Admin on Jul 18, 2022 11:47:26 GMT