Post by lisagalvan on Aug 3, 2022 16:21:28 GMT
Have 25,000 hours on T machines,so I'm having every issue a machine can have.But,when I suspect #10 CFS I go to debug screen #10 and look at my Hi ACFS voltage.I think it needs to be between 5 and 6.I have seen the CFS voltage at 7.2 and have had no complaints from staff.So what is a good indication it is definitely the CFS.Thank you.
|
|
|
Post by Guesttoday on Aug 3, 2022 17:27:04 GMT
If memory serves me correctly, CFS range should be 2 to 5. Anything above 6 is an indicator of a bad CFS. I would like to say that the 7.2 reading you have could be the result of the initial spike reading but I don't think I've come across a reading that high at "start up". It's usually like a 5.6 or below 6. I'd go into the TS guide to verify the ranges and follow steps to check. In your case, I would change out the CFS.
Also, is the CFS reading constantly on the high side or within that 2-5 range?
|
|
|
Post by lisagalvan on Aug 3, 2022 17:57:26 GMT
The intermittent problem is temp dropping and condo going high.Suspect a flow error causing it,but no one ever sees a flow error prior to these issues.Temp and condo will stabilize by the time I get to floor.Have ohmed all valves-ok.All 3 NTCs were replaced year ago,checked dear and flow brushes and pressures-ok.Dear motor is 10 years old,so may be faulty. But,I did wonder if CFS could be acting up.CFS spike is at 7.2 and I thought that was too high.I watch the CFS during tx and it is between 3 and 5.But it is intermittent issue so far.
|
|
|
Post by Biomed1 on Aug 6, 2022 2:59:11 GMT
Inspect balance chamber mechanicals to make sure valve plungers slide smoothly and that membrane isn't torn.
|
|
|
Post by lisagalvan on Aug 6, 2022 14:09:48 GMT
I did do the BC test to check if torn/leaking,but it passed.
|
|
|
Post by btrain on Sept 15, 2022 13:26:27 GMT
any news on this issue I am having somewhat of the same issue
|
|
bess
Junior Member
Posts: 59
|
Post by bess on Sept 15, 2022 14:22:23 GMT
Watch your flow rate in Debug screen during dialysis and see if it ever goes into eigentach - it's easier to watch that or flow error instead of CFS switching time. Also does it happen only during dialysis or does it happen during cleaning programs as well. If only during dialysis - is it random or does it occur with a filling program?
|
|