I don't often work on Fresenius machines anymore so I have a question that is probably easy for those who do. I have a T machine that boots all the way up but has a blank display. Even with the display blank if you hunt for the RINSE button you can still run it through that cycle, so I know the display is taking commands.
Is that a screen issue or is there some sort of graphics card or cable that could cause this? Thanks to all.
|
|
stuff
Full Member
1111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111
Posts: 240
|
Post by stuff on Aug 5, 2022 16:10:01 GMT
Sounds like the back light cable might be unplugged most likely from the screen side. Was the card cage worked in recently?
|
|
|
Post by Biomed 1 on Aug 6, 2022 2:29:57 GMT
Like stuff stated, most likely a loose cable on screen side.
|
|
pacnw
Full Member
Posts: 187
|
Post by pacnw on Aug 8, 2022 18:28:34 GMT
Check p-4 on the function board and make sure the wires from the 90 degree connector are up, not down. I'd also look at P41 for the same thing.
|
|
|
Post by vantech on Aug 20, 2022 9:05:45 GMT
I believe it is the J8 or J7 cable at the back of display connection. Had that happen to me a couple of times.
|
|