Post by texasbiomed on Aug 9, 2022 19:21:55 GMT
I have a high flow error when machine is connected to a bi-bag only. Troubleshooting guide isn't taking me anywhere. Anybody seen this before?
|
|
pacnw
Full Member
Posts: 187
|
Post by pacnw on Aug 9, 2022 21:14:49 GMT
check valve 100. while there also look at valve 103. any corrosion??? tighten the air separator cap if you find any.....
|
|
|
Post by texasbiomed on Aug 11, 2022 16:48:53 GMT
**Fix** Balancing Chamber diaphragms needed to be replaced.
|
|
|
Post by Easy Fixes Only on Aug 11, 2022 18:13:26 GMT
Thanks for posting your fix!
|
|
stuff
Full Member
1111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111111
Posts: 240
|
Post by stuff on Aug 12, 2022 18:49:11 GMT
How many hours on the machine?
|
|
|
Post by Diaphram on Aug 22, 2022 20:09:37 GMT
Starting to see this around the 20k mark on machines. I start to rebuild them during annuals as soon as i see a couple fail. how many hours did this machine have?
|
|