chase
New Member
Posts: 16
|
Post by chase on Aug 11, 2022 18:17:42 GMT
I plan to install a UPS for use with a Fresenius 2008K. I am wondering whether there is anyone here who has done it before. If so, what size UPS did you use?
|
|
|
UPS size
Aug 12, 2022 11:09:21 GMT
via mobile
Post by anon’s face on Aug 12, 2022 11:09:21 GMT
I plan to install a UPS for use with a Fresenius 2008K. I am wondering whether there is anyone here who has done it before. If so, what size UPS did you use? Why in the world?
|
|
|
Post by gfreely on Aug 12, 2022 11:46:48 GMT
I guess if a generator is not available?
Might want to talk directly to Fresenius about this one though.
|
|
|
Post by Chuck Weddle on Aug 12, 2022 12:53:35 GMT
I would think it would have to be HUGE and be capable of supplying pure sinewave. What would you use to power your RO?
|
|
chase
New Member
Posts: 16
|
Post by chase on Aug 13, 2022 17:52:40 GMT
I am thinking that the sudden voltage drop when power fails is probably not good for the 2008K power supply. A UPS would keep the voltage fed to the dialysis machine stable until the diesel generator kicks in. I do not have a solution for the RO at the moment. I am in the process of checking with the RO manufacturer.
|
|
|
Post by gnurk on Aug 14, 2022 12:29:06 GMT
most generator kick in in less than 50milisec which should not affect your machines, if it isnt talk to your generator people
|
|
chase
New Member
Posts: 16
|
Post by chase on Aug 14, 2022 17:39:51 GMT
I will check with the generator vendor. Thanks.
|
|
|
UPS size
Aug 18, 2022 11:20:34 GMT
via mobile
Post by anon’s face on Aug 18, 2022 11:20:34 GMT
most generator kick in in less than 50milisec which should not affect your machines, if it isnt talk to your generator people What generator kicks on that quickly!? Typically, a generator waits a second or two to verify power loss then the engine starts, then the transfer occurs. This is always enough in-center, and at home, to have your 2008 machine lose power and alarm. In my opinion, a UPS is a terrible idea and inviting trouble.
|
|