Post by gfreely on Aug 17, 2022 12:58:18 GMT
Have a 103 unit backflowing into the disinfect jug. Didn't start until after the priming port was upgraded (from the black to the opaque port) on the pump. Was TOLD the duck bill was installed but it is my understanding (and Gambro/MarCor/Evoqua/Whatever-the-hell-else-they-are-called-now agreed) that is probably the only thing that should cause that.
I am already scheduling to come in after hours to disassemble and verify this weekend but was curious if anyone had any drawing/pic/etc of these connections that they could share?
|
|
|
Post by gfreely on Sept 23, 2022 14:49:26 GMT
For posterity: turns out the little "guide" was not installed over the duck bill.
|
|
|
Post by Chuck Weddle on Sept 23, 2022 15:21:49 GMT
For posterity: turns out the little "guide" was not installed over the duck bill. Bill is a strange name for a duck!
|
|