Post by vantech on Aug 20, 2022 8:08:26 GMT
A fellow technician had a "Patient Alarm" in the top banner. I have looked in operators manual and it states that if continues, contact a qualified technician. Looked in the troubleshooting manual, I could not find it. I have been a technician for a long time and have 20 years experience on Fresenius machines. I have never seen this alarm. Any help would be appreciated.
|
|
|
Post by DRSGME on Jan 5, 2023 23:46:21 GMT
Currently dealing with this alarm on a 2008k2. Originally was pulled for leaking cause by a busted check valve. After replacing it, booted up the machine and got this alarm while trying to bring it up to conductivity. Machine has 14650 hrs. A biomed on a dialysis biomed sub reddit is the only place i found any sort of fix. He stated that he swapped the Power Logic Board and his issue was fixed. I tried that and alarm still shows up. I also tried swapping out the blood pump module as well as the venous module with known working modules and I still get the alarm. Any other suggestions here?
|
|
|
Post by gnurk on Jan 6, 2023 14:23:09 GMT
call tech services
|
|
|
Post by DRSGME on Jan 23, 2023 20:35:22 GMT
The Motherboard was the culprit in my "patient alarm"
|
|